Hello,
I have a problem with dispcaGUI.
My system:
MacBook Pro with Monitor U2713HM DELL
Yosemite
Color Munki PhotoAfter 1 hour metering the screen tells me:
“Computing update to calibration curves…
Black drift was nan DE
The instrument can be removed from the screen”I don´t understand what´s happening there.
What does this message mean?Regards, Markus
Calibrite ColorChecker Studio on Amazon
Disclosure: As an Amazon Associate I earn from qualifying purchases.2015-05-22 at 21:42 #591Hier der Anhang mit den Logfiles.
Attachments:
You must be logged in to view attached files.2015-05-22 at 21:56 #593Hi,
try disabling black level drift compensation.
2015-05-23 at 11:12 #594Hi!
Thank you!
I tried this but now there is another problem…2015-05-23 12:37:35,048 Patch 62 of 64 DE 0.054697, W.DE 0.054697, peqDE 0.076771, Repeat
2015-05-23 12:37:40,807 Patch 62 of 64 DE 0.046346, W.DE 25.595311, peqDE 0.068614, Repeat (got worse)
2015-05-23 12:37:46,568 Patch 62 of 64 DE 0.026241, W.DE 15.761124, peqDE 15.761195, Repeat
2015-05-23 12:37:52,332 Patch 62 of 64 DE 0.045740, W.DE 0.045740, W.peqDE 0.068614, Fail ( > 0.050000)
2015-05-23 12:37:58,177 Patch 63 of 64 DE 0.107010, W.DE 0.107010, peqDE 0.070682, Repeat
2015-05-23 12:38:22,173 Patch 63 of 64 DE 0.107010, W.DE 0.107010, peqDE 0.070682, Repeat
2015-05-23 12:38:28,355 Patch 63 of 64 DE 0.022374, W.DE 10.467839, peqDE 10.468054, Repeat
2015-05-23 12:38:34,121 Patch 63 of 64 DE 0.045021, W.DE 0.045021, peqDE 0.086982, Repeat
2015-05-23 12:38:39,917 Patch 63 of 64 DE 0.057709, W.DE 26.849092, peqDE 0.086982, Repeat (got worse)
2015-05-23 12:38:45,675 Patch 63 of 64 DE 0.036248, W.DE 29.163642, peqDE 0.086982, Repeat (got worse)
2015-05-23 12:38:51,434 Patch 63 of 64 DE 0.073884, W.DE 0.073884, peqDE 0.026817, Repeat (got worse)
2015-05-23 12:38:57,193 Patch 63 of 64 DE 0.093724, W.DE 81.953985, peqDE 0.086982, Repeat (got worse)
2015-05-23 12:39:02,975 Patch 63 of 64 DE 0.041675, W.DE 32.229442, peqDE 0.086982, Repeat (got worse)
2015-05-23 12:39:08,733 Patch 63 of 64 DE 0.039162, W.DE 32.766620, peqDE 0.086982, Repeat (got worse)
2015-05-23 12:39:14,492 Patch 63 of 64 DE 0.071839, W.DE 33.679686, peqDE 0.086982, Repeat (got worse)
2015-05-23 12:39:20,251 Patch 63 of 64 DE 0.045021, W.DE 0.045021, W.peqDE 0.086982, Fail ( > 0.050000)
2015-05-23 12:39:26,167 Patch 64 of 64 DE 0.007048, W.DE 0.007048, W.peqDE 0.007048, OK ( < 0.050000)
2015-05-23 12:39:26,168
2015-05-23 12:39:26,169 Brightness error = -0.033924 cd/m^2 (is 79.004316, should be 79.038239)
2015-05-23 12:39:26,170 White point error = 0.118328 deltaE
2015-05-23 12:39:26,171 Maximum neutral error (@ 0.502932) = 0.598883 deltaE
2015-05-23 12:39:26,171 Average neutral error = 0.283521 deltaE
2015-05-23 12:39:26,172 Failed to meet target 0.600000 delta E, got worst case 0.069684
2015-05-23 12:39:26,172 Number of measurements taken = 234
2015-05-23 12:39:26,173 Computing update to calibration curves…
2015-05-23 12:39:26,574 The instrument can be removed from the screen.I don´t understand the message…
Best regards, Markus
2015-05-23 at 13:23 #595There is no problem in the above log, in fact it seems to have achieved a good result (average neutral error 0.28 dE, max 0.6).
2015-05-23 at 14:28 #596?
And now? The screen tells me: “verbleibende Zeit: -..-“.
I can only click “Abbrechen” or “Pause”.
And where is my generated file?Regards, Markus
2015-05-23 at 14:30 #597So it hangs there? That’s strange as there is no error message. If you open up Mac OS X activity viewer, do you see the ‘dispcal’ process? If yes, that’s where it hangs. Try to end that process.
2015-05-23 at 14:32 #598I did it.
Now: “Es wurde keine Kalibrierungsdatei erstellt”.2015-05-23 at 14:33 #599Ok, please attach the full log of the run. Thanks.
2015-05-23 at 14:35 #600Here it is:
~~~~
2015-05-23 11:48:32,265 ——————————————————————————–
2015-05-23 11:48:32,265 Arbeitsverzeichnis:
2015-05-23 11:48:32,266 /
2015-05-23 11:48:32,266 var/
2015-05-23 11:48:32,267 folders/
2015-05-23 11:48:32,267 zc/
2015-05-23 11:48:32,267 x4kzk2k105b4t3dy3zqp5jg80000gn/
2015-05-23 11:48:32,268 T/
2015-05-23 11:48:32,268 dispcalGUI-1vQ3tq/
2015-05-23 11:48:32,268
2015-05-23 11:48:32,268 Befehlszeile:
2015-05-23 11:48:32,269 dispcal
2015-05-23 11:48:32,269 -v2
2015-05-23 11:48:32,270 -d2
2015-05-23 11:48:32,270 -c1
2015-05-23 11:48:32,270 -yl
2015-05-23 11:48:32,270 “-P0.500261096606,0.500628930818,1.5”
2015-05-23 11:48:32,271 -F
2015-05-23 11:48:32,271 -qm
2015-05-23 11:48:32,271 -t6500
2015-05-23 11:48:32,271 -b80.0
2015-05-23 11:48:32,272 -g2.2
2015-05-23 11:48:32,272 -f1.0
2015-05-23 11:48:32,272 -k0.0
2015-05-23 11:48:32,272 -A4.0
2015-05-23 11:48:32,273 -b
2015-05-23 11:48:32,273 “U2713HM 2015-05-23 80cdm² D6500 2.2 M-S XYZLUT+MTX”
2015-05-23 11:48:32,273
2015-05-23 11:48:32,287 dispcalGUI: Starting interaction with subprocess
2015-05-23 11:48:32,367 Setting up the instrument
2015-05-23 11:48:32,591 Instrument Type: ColorMunki
2015-05-23 11:48:32,591 Serial Number: 2069661
2015-05-23 11:48:32,592 Firmware version: 288
2015-05-23 11:48:32,592 Chip ID: 01-0473AE160000E5
2015-05-23 11:48:32,592 Version string: ‘colormunki FW V1.32 Build Nr. 1303’
2015-05-23 11:48:32,593 Calibration Ver.: 6
2015-05-23 11:48:32,593 Production No.: 117494
2015-05-23 11:48:32,796 Display type ignored – instrument doesn’t support display type selection
2015-05-23 11:48:32,809 dispcal: Warning – new_dispwin: frame buffer depth 8 != GammaTable depth 10
2015-05-23 11:48:32,810
2015-05-23 11:48:33,086 Set instrument sensor to calibration position,
2015-05-23 11:48:33,087 and then hit any key to continue,
2015-05-23 11:48:33,099 dispcalGUI: Waiting for send buffer
2015-05-23 11:51:22,751 dispcalGUI: Sending buffer: ‘ ‘
2015-05-23 11:51:22,804 or hit Esc or Q to abort:
2015-05-23 11:51:25,866 Calibration complete
2015-05-23 11:51:25,867
2015-05-23 11:51:25,867 Place instrument on test window.
2015-05-23 11:51:25,879 dispcalGUI: Waiting for send buffer
2015-05-23 11:52:14,211 dispcalGUI: Sending buffer: ‘ ‘
2015-05-23 11:52:14,266 Hit Esc or Q to give up, any other key to continue:
2015-05-23 11:52:14,529 Display type is ‘l’
2015-05-23 11:52:14,530 Target white = 6500.000000 degrees kelvin Daylight spectrum
2015-05-23 11:52:14,531 Target white brightness = 80.000000 cd/m^2
2015-05-23 11:52:14,531 Target black brightness = native brightness
2015-05-23 11:52:14,532 Black point device hack is enabled
2015-05-23 11:52:14,532 Target advertised gamma = 2.200000
2015-05-23 11:52:14,533
2015-05-23 11:52:14,534 Display adjustment menu:
2015-05-23 11:52:14,534 Press 1 .. 7
2015-05-23 11:52:14,534 1) Black level (CRT: Offset/Brightness)
2015-05-23 11:52:14,535 2) White point (Color temperature, R,G,B, Gain/Contrast)
2015-05-23 11:52:14,535 3) White level (CRT: Gain/Contrast, LCD: Brightness/Backlight)
2015-05-23 11:52:14,536 4) Black point (R,G,B, Offset/Brightness)
2015-05-23 11:52:14,536 5) Check all
2015-05-23 11:52:14,537 6) Measure and set ambient for viewing condition adjustment
2015-05-23 11:52:14,537 7) Continue on to calibration
2015-05-23 11:52:14,538 8) Exit
2015-05-23 11:52:14,550 dispcalGUI: Waiting for send buffer
2015-05-23 11:52:18,588 dispcalGUI: Sending buffer: ‘2’
2015-05-23 11:52:18,646 Doing some initial measurements
2015-05-23 11:52:21,276 Measured display update delay of 77 msec, using delay of 202 msec & 0 msec inst reaction
2015-05-23 11:52:32,720 Red = XYZ 31.861 16.283 0.630
2015-05-23 11:52:32,722 Green = XYZ 28.658 59.395 5.535
2015-05-23 11:52:32,722 Blue = XYZ 14.555 3.809 78.346
2015-05-23 11:52:35,848 White = XYZ 75.103 79.292 84.534
2015-05-23 11:52:35,863
2015-05-23 11:52:35,865 Adjust R,G & B gain to get target x,y. Press space when done.
2015-05-23 11:52:35,866 Target Br 80.00, x 0.3128 , y 0.3291
2015-05-23 12:03:48,559 dispcalGUI: Sending buffer: ‘ ‘
2015-05-23 12:03:48,751 / Current Br 79.11, x 0.3122+, y 0.3279+ DE 0.7 R+ G+ B–
2015-05-23 12:03:48,751
2015-05-23 12:03:48,752 Press 1 .. 7
2015-05-23 12:03:48,753 1) Black level (CRT: Offset/Brightness)
2015-05-23 12:03:48,753 2) White point (Color temperature, R,G,B, Gain/Contrast)
2015-05-23 12:03:48,753 3) White level (CRT: Gain/Contrast, LCD: Brightness/Backlight)
2015-05-23 12:03:48,754 4) Black point (R,G,B, Offset/Brightness)
2015-05-23 12:03:48,754 5) Check all
2015-05-23 12:03:48,755 6) Measure and set ambient for viewing condition adjustment
2015-05-23 12:03:48,756 7) Continue on to calibration
2015-05-23 12:03:48,756 8) Exit
2015-05-23 12:03:48,769 dispcalGUI: Waiting for send buffer
2015-05-23 12:03:49,190 dispcalGUI: Sending buffer: ‘2’
2015-05-23 12:03:49,246 Doing some initial measurements
2015-05-23 12:03:52,084 White = XYZ 75.326 79.118 86.827
2015-05-23 12:03:52,097
2015-05-23 12:03:52,098 Adjust R,G & B gain to get target x,y. Press space when done.
2015-05-23 12:03:52,099 Target Br 80.00, x 0.3128 , y 0.3291
2015-05-23 12:03:54,934 dispcalGUI: Sending buffer: ‘ ‘
2015-05-23 12:03:55,130 / Current Br 79.11, x 0.3122+, y 0.3279+ DE 0.7 R+ G+ B–
2015-05-23 12:03:55,131
2015-05-23 12:03:55,131 Press 1 .. 7
2015-05-23 12:03:55,132 1) Black level (CRT: Offset/Brightness)
2015-05-23 12:03:55,133 2) White point (Color temperature, R,G,B, Gain/Contrast)
2015-05-23 12:03:55,133 3) White level (CRT: Gain/Contrast, LCD: Brightness/Backlight)
2015-05-23 12:03:55,133 4) Black point (R,G,B, Offset/Brightness)
2015-05-23 12:03:55,134 5) Check all
2015-05-23 12:03:55,134 6) Measure and set ambient for viewing condition adjustment
2015-05-23 12:03:55,135 7) Continue on to calibration
2015-05-23 12:03:55,136 8) Exit
2015-05-23 12:03:55,147 dispcalGUI: Waiting for send buffer
2015-05-23 12:03:56,931 dispcalGUI: Sending buffer: ‘7’
2015-05-23 12:03:56,986 Commencing display calibration
2015-05-23 12:04:40,428 Patch 9 of 9
2015-05-23 12:04:40,429 Black = XYZ 0.0787 0.0757 0.1446
2015-05-23 12:04:40,430 Red = XYZ 31.771 16.246 0.649
2015-05-23 12:04:40,431 Green = XYZ 28.545 59.169 5.539
2015-05-23 12:04:40,431 Blue = XYZ 15.012 3.916 80.763
2015-05-23 12:04:40,432 White = XYZ 75.309 79.096 86.807
2015-05-23 12:09:55,981 Patch 64 of 64
2015-05-23 12:09:56,122 Had to scale brightness from 80.000000 to 79.038239 to fit within gamut,
2015-05-23 12:09:56,124 corresponding to aprox. RGB 0.999238 1.000000 0.994678
2015-05-23 12:09:56,125 Target white value is XYZ 75.124134 79.038239 86.024576 [xy 0.312774 0.329070]
2015-05-23 12:09:56,125 Adjusted target black XYZ 0.0787 0.0757 0.1446, Lab 0.866 0.349 -1.126
2015-05-23 12:09:56,126 Target black after min adjust: XYZ 0.0787 0.0757 0.1446, Lab 0.866 0.349 -1.126
2015-05-23 12:09:56,140 Gamma curve input offset = 0.000000, output offset = 0.000958, power = 2.204983
2015-05-23 12:09:56,141 Total Iteration 3, Final Samples = 64 Final Repeat threshold = 0.600000
2015-05-23 12:09:56,141 Creating initial calibration curves…
2015-05-23 12:09:56,343 Doing iteration 1 with 16 sample points and repeat threshold of 1.200000 DE
2015-05-23 12:09:59,385 Patch 1 of 16 DE 0.170702, W.DE 0.170702, OK ( < 1.200000)
2015-05-23 12:10:03,261 Patch 2 of 16 DE 3.457432, W.DE 3.457432, Repeat
2015-05-23 12:10:06,743 Patch 2 of 16 DE 0.823823, W.DE 0.823823, OK ( < 1.200000)
2015-05-23 12:10:10,575 Patch 3 of 16 DE 2.634584, W.DE 2.634584, Repeat
2015-05-23 12:10:13,986 Patch 3 of 16 DE 0.612270, W.DE 0.612270, OK ( < 1.200000)
2015-05-23 12:10:17,817 Patch 4 of 16 DE 2.152030, W.DE 2.152030, Repeat
2015-05-23 12:10:21,350 Patch 4 of 16 DE 0.503031, W.DE 0.503031, OK ( < 1.200000)
2015-05-23 12:10:25,169 Patch 5 of 16 DE 2.597595, W.DE 2.597595, Repeat
2015-05-23 12:10:28,794 Patch 5 of 16 DE 0.456418, W.DE 0.456418, OK ( < 1.200000)
2015-05-23 12:10:32,598 Patch 6 of 16 DE 1.879182, W.DE 1.879182, Repeat
2015-05-23 12:10:36,151 Patch 6 of 16 DE 0.254462, W.DE 0.254462, OK ( < 1.200000)
2015-05-23 12:10:39,962 Patch 7 of 16 DE 1.281314, W.DE 1.281314, Repeat
2015-05-23 12:10:43,523 Patch 7 of 16 DE 0.572668, W.DE 0.572668, OK ( < 1.200000)
2015-05-23 12:10:47,497 Patch 8 of 16 DE 0.932680, W.DE 0.932680, OK ( < 1.200000)
2015-05-23 12:10:52,843 Patch 9 of 16 DE 0.663133, W.DE 0.663133, OK ( < 1.200000)
2015-05-23 12:10:59,132 Patch 10 of 16 DE 0.608742, W.DE 0.608742, OK ( < 1.200000)
2015-05-23 12:11:05,429 Patch 11 of 16 DE 1.765243, W.DE 1.765243, Repeat
2015-05-23 12:11:11,303 Patch 11 of 16 DE 0.566211, W.DE 0.566211, OK ( < 1.200000)
2015-05-23 12:11:17,641 Patch 12 of 16 DE 2.854170, W.DE 2.854170, Repeat
2015-05-23 12:11:23,559 Patch 12 of 16 DE 1.152130, W.DE 1.152130, OK ( < 1.200000)
2015-05-23 12:11:29,891 Patch 13 of 16 DE 2.015992, W.DE 2.015992, Repeat
2015-05-23 12:11:35,815 Patch 13 of 16 DE 0.931235, W.DE 0.931235, OK ( < 1.176294)
2015-05-23 12:11:42,075 Patch 14 of 16 DE 0.559671, W.DE 0.559671, Repeat
2015-05-23 12:12:06,217 Patch 14 of 16 DE 0.555975, W.DE 0.555975, Repeat
2015-05-23 12:12:12,397 Patch 14 of 16 DE 0.270860, W.DE 0.270860, OK ( < 0.342094)
2015-05-23 12:12:18,565 Patch 15 of 16 DE 0.135902, W.DE 0.135902, Repeat
2015-05-23 12:12:24,323 Patch 15 of 16 DE 0.103664, W.DE 0.103664, Repeat
2015-05-23 12:12:48,334 Patch 15 of 16 DE 0.102197, W.DE 0.102197, Repeat
2015-05-23 12:12:54,523 Patch 15 of 16 DE 0.035562, W.DE 0.035562, OK ( < 0.050000)
2015-05-23 12:13:00,468 Patch 16 of 16 DE 0.016752, W.DE 16.752264, Repeat
2015-05-23 12:13:24,465 Patch 16 of 16 DE 0.016752, W.DE 16.752264, Repeat
2015-05-23 12:13:30,655 Patch 16 of 16 DE 0.043805, W.DE 43.805118, Repeat (got worse)
2015-05-23 12:13:36,470 Patch 16 of 16 DE 0.004263, W.DE 0.004263, OK ( < 0.050000)
2015-05-23 12:13:36,472
2015-05-23 12:13:36,472 Brightness error = 0.013880 cd/m^2 (is 79.052119, should be 79.038239)
2015-05-23 12:13:36,473 White point error = 0.170557 deltaE
2015-05-23 12:13:36,473 Maximum neutral error (@ 0.120656) = 1.166136 deltaE
2015-05-23 12:13:36,474 Average neutral error = 0.536238 deltaE
2015-05-23 12:13:36,474 Number of measurements taken = 42
2015-05-23 12:13:36,475 Computing update to calibration curves…
2015-05-23 12:13:36,620 Doing iteration 2 with 32 sample points and repeat threshold of 0.848528 DE
2015-05-23 12:13:39,753 Patch 1 of 32 DE 0.143478, W.DE 0.143478, OK ( < 0.848528)
2015-05-23 12:13:43,297 Patch 2 of 32 DE 1.974595, W.DE 1.974595, Repeat
2015-05-23 12:13:46,665 Patch 2 of 32 DE 0.255043, W.DE 0.255043, OK ( < 0.848528)
2015-05-23 12:13:50,425 Patch 3 of 32 DE 0.273926, W.DE 0.273926, OK ( < 0.848528)
2015-05-23 12:13:54,178 Patch 4 of 32 DE 0.759122, W.DE 0.759122, OK ( < 0.848528)
2015-05-23 12:13:57,918 Patch 5 of 32 DE 0.606369, W.DE 0.606369, OK ( < 0.848528)
2015-05-23 12:14:01,665 Patch 6 of 32 DE 0.794763, W.DE 0.794763, OK ( < 0.848528)
2015-05-23 12:14:05,419 Patch 7 of 32 DE 0.333253, W.DE 0.333253, OK ( < 0.848528)
2015-05-23 12:14:09,173 Patch 8 of 32 DE 0.268661, W.DE 0.268661, OK ( < 0.848528)
2015-05-23 12:14:12,927 Patch 9 of 32 DE 0.569938, W.DE 0.569938, OK ( < 0.848528)
2015-05-23 12:14:16,668 Patch 10 of 32 DE 0.445874, W.DE 0.445874, OK ( < 0.848528)
2015-05-23 12:14:20,406 Patch 11 of 32 DE 0.596662, W.DE 0.596662, OK ( < 0.848528)
2015-05-23 12:14:24,139 Patch 12 of 32 DE 0.394251, W.DE 0.394251, OK ( < 0.848528)
2015-05-23 12:14:27,878 Patch 13 of 32 DE 0.530522, W.DE 0.530522, OK ( < 0.848528)
2015-05-23 12:14:31,611 Patch 14 of 32 DE 0.660931, W.DE 0.660931, OK ( < 0.848528)
2015-05-23 12:14:35,344 Patch 15 of 32 DE 0.628074, W.DE 0.628074, OK ( < 0.848528)
2015-05-23 12:14:39,540 Patch 16 of 32 DE 1.184320, W.DE 1.184320, Repeat
2015-05-23 12:14:43,324 Patch 16 of 32 DE 0.260743, W.DE 0.260743, OK ( < 0.848528)
2015-05-23 12:14:48,198 Patch 17 of 32 DE 1.326064, W.DE 1.326064, Repeat
2015-05-23 12:14:52,601 Patch 17 of 32 DE 0.509743, W.DE 0.509743, OK ( < 0.848528)
2015-05-23 12:14:58,345 Patch 18 of 32 DE 0.727132, W.DE 0.727132, OK ( < 0.848528)
2015-05-23 12:15:04,548 Patch 19 of 32 DE 1.179506, W.DE 1.179506, Repeat
2015-05-23 12:15:10,421 Patch 19 of 32 DE 0.554640, W.DE 0.554640, OK ( < 0.848528)
2015-05-23 12:15:16,654 Patch 20 of 32 DE 1.208292, W.DE 1.208292, Repeat
2015-05-23 12:15:22,520 Patch 20 of 32 DE 0.387689, W.DE 0.387689, OK ( < 0.848528)
2015-05-23 12:15:28,739 Patch 21 of 32 DE 1.586579, W.DE 1.586579, Repeat
2015-05-23 12:15:34,619 Patch 21 of 32 DE 0.597418, W.DE 0.597418, OK ( < 0.848528)
2015-05-23 12:15:40,851 Patch 22 of 32 DE 0.932864, W.DE 0.932864, Repeat
2015-05-23 12:15:46,718 Patch 22 of 32 DE 0.744154, W.DE 0.744154, OK ( < 0.848528)
2015-05-23 12:15:52,950 Patch 23 of 32 DE 0.953758, W.DE 0.953758, Repeat
2015-05-23 12:15:58,816 Patch 23 of 32 DE 0.676123, W.DE 0.676123, OK ( < 0.848528)
2015-05-23 12:16:05,041 Patch 24 of 32 DE 1.972389, W.DE 1.972389, Repeat
2015-05-23 12:16:10,928 Patch 24 of 32 DE 1.085499, W.DE 1.085499, Repeat
2015-05-23 12:16:35,391 Patch 24 of 32 DE 0.927411, W.DE 0.927411, Repeat
2015-05-23 12:16:41,652 Patch 24 of 32 DE 0.595437, W.DE 0.595437, OK ( < 0.848528)
2015-05-23 12:16:47,898 Patch 25 of 32 DE 1.546969, W.DE 1.546969, Repeat
2015-05-23 12:16:53,794 Patch 25 of 32 DE 0.197887, W.DE 0.197887, OK ( < 0.848528)
2015-05-23 12:17:00,026 Patch 26 of 32 DE 1.149297, W.DE 1.149297, Repeat
2015-05-23 12:17:05,920 Patch 26 of 32 DE 0.624866, W.DE 0.624866, OK ( < 0.653153)
2015-05-23 12:17:12,131 Patch 27 of 32 DE 0.642530, W.DE 0.642530, Repeat
2015-05-23 12:17:17,990 Patch 27 of 32 DE 0.379030, W.DE 0.379030, OK ( < 0.395157)
2015-05-23 12:17:24,136 Patch 28 of 32 DE 0.364006, W.DE 0.364006, Repeat
2015-05-23 12:17:29,967 Patch 28 of 32 DE 0.192243, W.DE 0.192243, OK ( < 0.211857)
2015-05-23 12:17:36,077 Patch 29 of 32 DE 0.122388, W.DE 0.122388, Repeat
2015-05-23 12:17:41,873 Patch 29 of 32 DE 0.078537, W.DE 0.078537, OK ( < 0.094301)
2015-05-23 12:17:47,939 Patch 30 of 32 DE 0.084593, W.DE 0.084593, Repeat
2015-05-23 12:18:11,950 Patch 30 of 32 DE 0.084593, W.DE 0.084593, Repeat
2015-05-23 12:18:18,139 Patch 30 of 32 DE 0.172792, W.DE 1.077067, Repeat (got worse)
2015-05-23 12:18:23,899 Patch 30 of 32 DE 0.051300, W.DE 0.051300, Repeat
2015-05-23 12:18:29,664 Patch 30 of 32 DE 0.075562, W.DE 0.075562, Repeat (got worse)
2015-05-23 12:18:35,423 Patch 30 of 32 DE 0.081191, W.DE 4.868123, Repeat (got worse)
2015-05-23 12:18:41,182 Patch 30 of 32 DE 0.145456, W.DE 1.918942, Repeat (got worse)
2015-05-23 12:18:46,940 Patch 30 of 32 DE 0.090036, W.DE 6.371217, Repeat (got worse)
2015-05-23 12:18:52,708 Patch 30 of 32 DE 0.079848, W.DE 0.079848, Repeat (got worse)
2015-05-23 12:18:58,475 Patch 30 of 32 DE 0.105495, W.DE 10.577041, Repeat (got worse)
2015-05-23 12:19:04,233 Patch 30 of 32 DE 0.087922, W.DE 0.087922, Repeat (got worse)
2015-05-23 12:19:09,992 Patch 30 of 32 DE 0.026475, W.DE 0.026475, OK ( < 0.050000)
2015-05-23 12:19:16,016 Patch 31 of 32 DE 0.082388, W.DE 47.997297, Repeat
2015-05-23 12:19:21,825 Patch 31 of 32 DE 0.031472, W.DE 0.031472, OK ( < 0.050000)
2015-05-23 12:19:27,720 Patch 32 of 32 DE 0.019021, W.DE 0.019021, OK ( < 0.050000)
2015-05-23 12:19:27,721
2015-05-23 12:19:27,722 Brightness error = -0.014752 cd/m^2 (is 79.023487, should be 79.038239)
2015-05-23 12:19:27,722 White point error = 0.143305 deltaE
2015-05-23 12:19:27,723 Maximum neutral error (@ 0.754707) = 0.794773 deltaE
2015-05-23 12:19:27,723 Average neutral error = 0.433227 deltaE
2015-05-23 12:19:27,724 Number of measurements taken = 66
2015-05-23 12:19:27,724 Computing update to calibration curves…
2015-05-23 12:19:27,924 Doing iteration 3 with 64 sample points and repeat threshold of 0.600000 DE
2015-05-23 12:19:31,062 Patch 1 of 64 DE 0.119470, W.DE 0.119470, W.peqDE 0.119470, OK ( < 0.600000)
2015-05-23 12:19:34,423 Patch 2 of 64 DE 2.700641, W.DE 2.700641, peqDE 2.610297, Repeat
2015-05-23 12:19:37,738 Patch 2 of 64 DE 0.971293, W.DE 0.971293, peqDE 1.059911, Repeat
2015-05-23 12:19:40,778 Patch 2 of 64 DE 0.203411, W.DE 0.203411, W.peqDE 0.134597, OK ( < 0.600000)
2015-05-23 12:19:44,228 Patch 3 of 64 DE 0.427387, W.DE 0.427387, W.peqDE 0.280529, OK ( < 0.600000)
2015-05-23 12:19:47,801 Patch 4 of 64 DE 0.183194, W.DE 0.183194, W.peqDE 0.514523, OK ( < 0.600000)
2015-05-23 12:19:51,467 Patch 5 of 64 DE 0.836286, W.DE 0.836286, peqDE 0.735862, Repeat
2015-05-23 12:19:54,926 Patch 5 of 64 DE 0.305972, W.DE 0.305972, W.peqDE 0.398288, OK ( < 0.600000)
2015-05-23 12:19:58,600 Patch 6 of 64 DE 1.081563, W.DE 1.081563, peqDE 1.333067, Repeat
2015-05-23 12:20:01,960 Patch 6 of 64 DE 0.796534, W.DE 0.796534, peqDE 0.680895, Repeat
2015-05-23 12:20:05,420 Patch 6 of 64 DE 0.171981, W.DE 0.171981, W.peqDE 0.403362, OK ( < 0.600000)
2015-05-23 12:20:09,095 Patch 7 of 64 DE 0.862618, W.DE 0.862618, peqDE 0.852038, Repeat
2015-05-23 12:20:12,455 Patch 7 of 64 DE 0.485444, W.DE 0.485444, W.peqDE 0.577867, OK ( < 0.600000)
2015-05-23 12:20:16,146 Patch 8 of 64 DE 0.830206, W.DE 0.830206, peqDE 1.179046, Repeat
2015-05-23 12:20:19,511 Patch 8 of 64 DE 0.181587, W.DE 0.181587, W.peqDE 0.488029, OK ( < 0.600000)
2015-05-23 12:20:23,201 Patch 9 of 64 DE 0.531911, W.DE 0.531911, W.peqDE 0.520126, OK ( < 0.600000)
2015-05-23 12:20:26,862 Patch 10 of 64 DE 0.631161, W.DE 0.631161, peqDE 0.610518, Repeat
2015-05-23 12:20:30,222 Patch 10 of 64 DE 0.247305, W.DE 0.247305, W.peqDE 0.292250, OK ( < 0.600000)
2015-05-23 12:20:33,911 Patch 11 of 64 DE 0.802750, W.DE 0.802750, peqDE 0.734823, Repeat
2015-05-23 12:20:37,286 Patch 11 of 64 DE 0.307144, W.DE 0.307144, W.peqDE 0.405989, OK ( < 0.600000)
2015-05-23 12:20:40,982 Patch 12 of 64 DE 0.639891, W.DE 0.639891, peqDE 0.493817, Repeat
2015-05-23 12:20:44,334 Patch 12 of 64 DE 0.213112, W.DE 0.213112, W.peqDE 0.413390, OK ( < 0.600000)
2015-05-23 12:20:48,024 Patch 13 of 64 DE 0.449662, W.DE 0.449662, W.peqDE 0.482545, OK ( < 0.600000)
2015-05-23 12:20:51,699 Patch 14 of 64 DE 0.260003, W.DE 0.260003, W.peqDE 0.255294, OK ( < 0.600000)
2015-05-23 12:20:55,374 Patch 15 of 64 DE 0.384528, W.DE 0.384528, W.peqDE 0.285031, OK ( < 0.600000)
2015-05-23 12:20:59,051 Patch 16 of 64 DE 0.268100, W.DE 0.268100, W.peqDE 0.501499, OK ( < 0.600000)
2015-05-23 12:21:02,725 Patch 17 of 64 DE 0.215515, W.DE 0.215515, W.peqDE 0.349019, OK ( < 0.600000)
2015-05-23 12:21:06,392 Patch 18 of 64 DE 0.315996, W.DE 0.315996, W.peqDE 0.407215, OK ( < 0.600000)
2015-05-23 12:21:10,054 Patch 19 of 64 DE 1.053830, W.DE 1.053830, peqDE 0.888237, Repeat
2015-05-23 12:21:13,499 Patch 19 of 64 DE 0.369434, W.DE 0.369434, W.peqDE 0.580124, OK ( < 0.600000)
2015-05-23 12:21:17,145 Patch 20 of 64 DE 0.759956, W.DE 0.759956, peqDE 0.819793, Repeat
2015-05-23 12:21:31,675 Patch 20 of 64 DE 0.708397, W.DE 0.708397, peqDE 1.001729, Repeat
2015-05-23 12:21:35,113 Patch 20 of 64 DE 0.169180, W.DE 0.169180, W.peqDE 0.501183, OK ( < 0.600000)
2015-05-23 12:21:38,781 Patch 21 of 64 DE 0.434862, W.DE 0.434862, W.peqDE 0.401733, OK ( < 0.600000)
2015-05-23 12:21:42,443 Patch 22 of 64 DE 0.688156, W.DE 0.688156, peqDE 0.718305, Repeat
2015-05-23 12:21:45,801 Patch 22 of 64 DE 0.256342, W.DE 0.256342, W.peqDE 0.352993, OK ( < 0.600000)
2015-05-23 12:21:49,490 Patch 23 of 64 DE 0.598234, W.DE 0.598234, W.peqDE 0.442945, OK ( < 0.600000)
2015-05-23 12:21:53,144 Patch 24 of 64 DE 0.651189, W.DE 0.651189, peqDE 0.348286, Repeat
2015-05-23 12:22:07,386 Patch 24 of 64 DE 0.651189, W.DE 0.651189, peqDE 0.348286, Repeat
2015-05-23 12:22:11,111 Patch 24 of 64 DE 0.710848, W.DE 0.710848, peqDE 1.194727, Repeat
2015-05-23 12:22:14,428 Patch 24 of 64 DE 0.133771, W.DE 0.133771, W.peqDE 0.571754, OK ( < 0.600000)
2015-05-23 12:22:18,089 Patch 25 of 64 DE 0.440131, W.DE 0.440131, W.peqDE 0.340008, OK ( < 0.600000)
2015-05-23 12:22:21,742 Patch 26 of 64 DE 0.622116, W.DE 0.622116, peqDE 0.334838, Repeat
2015-05-23 12:22:35,985 Patch 26 of 64 DE 0.622116, W.DE 0.622116, peqDE 0.334838, Repeat
2015-05-23 12:22:39,710 Patch 26 of 64 DE 0.618085, W.DE 0.618085, peqDE 1.033062, Repeat
2015-05-23 12:22:43,048 Patch 26 of 64 DE 0.523521, W.DE 0.523521, W.peqDE 0.195122, OK ( < 0.600000)
2015-05-23 12:22:46,716 Patch 27 of 64 DE 1.167574, W.DE 1.167574, peqDE 0.708789, Repeat
2015-05-23 12:22:50,062 Patch 27 of 64 DE 0.409121, W.DE 0.409121, peqDE 0.762669, Repeat
2015-05-23 12:23:04,368 Patch 27 of 64 DE 0.409121, W.DE 0.409121, peqDE 0.762669, Repeat
2015-05-23 12:23:08,094 Patch 27 of 64 DE 0.609964, W.DE 0.609964, peqDE 1.082002, Repeat (got worse)
2015-05-23 12:23:11,418 Patch 27 of 64 DE 0.309162, W.DE 0.309162, W.peqDE 0.281304, OK ( < 0.600000)
2015-05-23 12:23:15,064 Patch 28 of 64 DE 0.399115, W.DE 0.399115, peqDE 0.622570, Repeat
2015-05-23 12:23:18,409 Patch 28 of 64 DE 0.280022, W.DE 0.280022, W.peqDE 0.418685, OK ( < 0.600000)
2015-05-23 12:23:22,069 Patch 29 of 64 DE 0.626585, W.DE 0.626585, peqDE 0.766625, Repeat
2015-05-23 12:23:25,415 Patch 29 of 64 DE 0.404848, W.DE 0.404848, W.peqDE 0.170370, OK ( < 0.600000)
2015-05-23 12:23:29,164 Patch 30 of 64 DE 0.594251, W.DE 0.594251, peqDE 0.863964, Repeat
2015-05-23 12:23:43,498 Patch 30 of 64 DE 0.620126, W.DE 0.620126, peqDE 0.670066, Repeat
2015-05-23 12:23:47,272 Patch 30 of 64 DE 0.312858, W.DE 0.312858, W.peqDE 0.594224, OK ( < 0.600000)
2015-05-23 12:23:51,244 Patch 31 of 64 DE 0.467033, W.DE 0.467033, W.peqDE 0.435742, OK ( < 0.600000)
2015-05-23 12:23:55,430 Patch 32 of 64 DE 0.933377, W.DE 0.933377, peqDE 0.954277, Repeat
2015-05-23 12:24:10,702 Patch 32 of 64 DE 0.775165, W.DE 0.775165, peqDE 0.871910, Repeat
2015-05-23 12:24:14,977 Patch 32 of 64 DE 0.273965, W.DE 0.273965, W.peqDE 0.571106, OK ( < 0.600000)
2015-05-23 12:24:19,494 Patch 33 of 64 DE 0.919425, W.DE 0.919425, peqDE 0.768554, Repeat
2015-05-23 12:24:23,671 Patch 33 of 64 DE 0.168784, W.DE 0.168784, W.peqDE 0.168419, OK ( < 0.600000)
2015-05-23 12:24:28,509 Patch 34 of 64 DE 0.225747, W.DE 0.225747, W.peqDE 0.263827, OK ( < 0.600000)
2015-05-23 12:24:33,712 Patch 35 of 64 DE 0.314279, W.DE 0.314279, W.peqDE 0.331727, OK ( < 0.600000)
2015-05-23 12:24:39,379 Patch 36 of 64 DE 0.825496, W.DE 0.825496, peqDE 0.752561, Repeat
2015-05-23 12:24:44,703 Patch 36 of 64 DE 0.499409, W.DE 0.499409, W.peqDE 0.476572, OK ( < 0.600000)
2015-05-23 12:24:50,843 Patch 37 of 64 DE 0.264925, W.DE 0.264925, W.peqDE 0.479205, OK ( < 0.600000)
2015-05-23 12:24:56,981 Patch 38 of 64 DE 0.660939, W.DE 0.660939, peqDE 0.658111, Repeat
2015-05-23 12:25:18,996 Patch 38 of 64 DE 0.660939, W.DE 0.660939, peqDE 0.658111, Repeat
2015-05-23 12:25:25,235 Patch 38 of 64 DE 1.275493, W.DE 1.275493, peqDE 1.280370, Repeat (got worse)
2015-05-23 12:25:31,094 Patch 38 of 64 DE 0.658025, W.DE 0.658025, peqDE 0.654863, Repeat
2015-05-23 12:25:36,911 Patch 38 of 64 DE 0.555499, W.DE 0.555499, W.peqDE 0.509653, OK ( < 0.600000)
2015-05-23 12:25:43,057 Patch 39 of 64 DE 0.309365, W.DE 0.309365, W.peqDE 0.520938, OK ( < 0.600000)
2015-05-23 12:25:49,182 Patch 40 of 64 DE 0.377588, W.DE 0.377588, W.peqDE 0.370732, OK ( < 0.600000)
2015-05-23 12:25:55,306 Patch 41 of 64 DE 0.429527, W.DE 0.429527, W.peqDE 0.073557, OK ( < 0.600000)
2015-05-23 12:26:01,424 Patch 42 of 64 DE 0.614363, W.DE 0.614363, peqDE 0.356325, Repeat
2015-05-23 12:26:07,240 Patch 42 of 64 DE 0.300491, W.DE 0.300491, W.peqDE 0.599428, OK ( < 0.600000)
2015-05-23 12:26:13,379 Patch 43 of 64 DE 0.589445, W.DE 0.589445, peqDE 0.793768, Repeat
2015-05-23 12:26:19,195 Patch 43 of 64 DE 0.345122, W.DE 0.345122, W.peqDE 0.112921, OK ( < 0.600000)
2015-05-23 12:26:25,312 Patch 44 of 64 DE 0.502615, W.DE 0.502615, peqDE 0.761765, Repeat
2015-05-23 12:26:49,674 Patch 44 of 64 DE 0.444526, W.DE 0.444526, W.peqDE 0.336233, OK ( < 0.600000)
2015-05-23 12:26:55,956 Patch 45 of 64 DE 0.518277, W.DE 0.518277, peqDE 0.790200, Repeat
2015-05-23 12:27:20,369 Patch 45 of 64 DE 0.538081, W.DE 0.538081, W.peqDE 0.270270, OK ( < 0.600000)
2015-05-23 12:27:26,665 Patch 46 of 64 DE 0.245220, W.DE 0.245220, W.peqDE 0.398927, OK ( < 0.600000)
2015-05-23 12:27:32,783 Patch 47 of 64 DE 0.982431, W.DE 0.982431, peqDE 1.005046, Repeat
2015-05-23 12:27:38,750 Patch 47 of 64 DE 0.633100, W.DE 0.633100, peqDE 0.661373, Repeat
2015-05-23 12:28:03,047 Patch 47 of 64 DE 0.518020, W.DE 0.518020, W.peqDE 0.491521, OK ( < 0.600000)
2015-05-23 12:28:09,337 Patch 48 of 64 DE 0.551213, W.DE 0.551213, peqDE 0.946294, Repeat
2015-05-23 12:28:33,655 Patch 48 of 64 DE 0.526550, W.DE 0.526550, W.peqDE 0.224700, OK ( < 0.600000)
2015-05-23 12:28:39,953 Patch 49 of 64 DE 0.548601, W.DE 0.548601, peqDE 0.722318, Repeat
2015-05-23 12:28:45,761 Patch 49 of 64 DE 0.067593, W.DE 0.067593, W.peqDE 0.417593, OK ( < 0.600000)
2015-05-23 12:28:51,893 Patch 50 of 64 DE 0.730903, W.DE 0.730903, peqDE 0.744429, Repeat
2015-05-23 12:28:57,745 Patch 50 of 64 DE 0.159512, W.DE 0.159512, W.peqDE 0.162849, OK ( < 0.515258)
2015-05-23 12:29:03,892 Patch 51 of 64 DE 0.533504, W.DE 0.533504, peqDE 0.520590, Repeat
2015-05-23 12:29:28,211 Patch 51 of 64 DE 0.533504, W.DE 0.533504, peqDE 0.520590, Repeat
2015-05-23 12:29:34,492 Patch 51 of 64 DE 0.644451, W.DE 0.644451, peqDE 0.643177, Repeat
2015-05-23 12:29:40,331 Patch 51 of 64 DE 0.726015, W.DE 0.726015, peqDE 0.708672, Repeat
2015-05-23 12:29:46,204 Patch 51 of 64 DE 0.228943, W.DE 0.228943, W.peqDE 0.210172, OK ( < 0.424352)
2015-05-23 12:29:52,350 Patch 52 of 64 DE 0.701950, W.DE 0.701950, peqDE 0.666741, Repeat
2015-05-23 12:29:58,217 Patch 52 of 64 DE 0.258760, W.DE 0.258760, W.peqDE 0.227574, OK ( < 0.343927)
2015-05-23 12:30:04,355 Patch 53 of 64 DE 0.421855, W.DE 0.421855, peqDE 0.455361, Repeat
2015-05-23 12:30:10,207 Patch 53 of 64 DE 0.149199, W.DE 0.149199, W.peqDE 0.256835, OK ( < 0.273537)
2015-05-23 12:30:16,317 Patch 54 of 64 DE 0.275566, W.DE 0.275566, peqDE 0.287173, Repeat
2015-05-23 12:30:22,166 Patch 54 of 64 DE 0.153902, W.DE 0.153902, W.peqDE 0.124533, OK ( < 0.212717)
2015-05-23 12:30:28,244 Patch 55 of 64 DE 0.222058, W.DE 0.222058, peqDE 0.238012, Repeat
2015-05-23 12:30:52,406 Patch 55 of 64 DE 0.215097, W.DE 0.215097, peqDE 0.159499, Repeat
2015-05-23 12:30:58,623 Patch 55 of 64 DE 0.252727, W.DE 0.252727, peqDE 0.158198, Repeat (got worse)
2015-05-23 12:31:04,404 Patch 55 of 64 DE 0.081680, W.DE 0.081680, W.peqDE 0.062486, OK ( < 0.160985)
2015-05-23 12:31:10,471 Patch 56 of 64 DE 0.159700, W.DE 0.159700, peqDE 0.195532, Repeat
2015-05-23 12:31:34,560 Patch 56 of 64 DE 0.165896, W.DE 0.165896, peqDE 0.147633, Repeat
2015-05-23 12:31:40,772 Patch 56 of 64 DE 0.110291, W.DE 0.110291, peqDE 0.138383, Repeat
2015-05-23 12:31:46,537 Patch 56 of 64 DE 0.060020, W.DE 0.060020, W.peqDE 0.076319, OK ( < 0.117839)
2015-05-23 12:31:52,605 Patch 57 of 64 DE 0.183952, W.DE 0.183952, peqDE 0.135552, Repeat
2015-05-23 12:32:16,645 Patch 57 of 64 DE 0.183952, W.DE 0.183952, peqDE 0.135552, Repeat
2015-05-23 12:32:22,847 Patch 57 of 64 DE 0.254901, W.DE 0.254901, peqDE 0.286041, Repeat (got worse)
2015-05-23 12:32:28,650 Patch 57 of 64 DE 0.069575, W.DE 0.069575, W.peqDE 0.018900, OK ( < 0.082754)
2015-05-23 12:32:34,681 Patch 58 of 64 DE 0.078373, W.DE 0.078373, peqDE 0.146513, Repeat
2015-05-23 12:32:58,706 Patch 58 of 64 DE 0.078373, W.DE 0.078373, peqDE 0.146513, Repeat
2015-05-23 12:33:04,903 Patch 58 of 64 DE 0.105668, W.DE 0.105668, peqDE 0.163426, Repeat (got worse)
2015-05-23 12:33:10,661 Patch 58 of 64 DE 0.110322, W.DE 0.110322, peqDE 0.179390, Repeat (got worse)
2015-05-23 12:33:16,420 Patch 58 of 64 DE 0.058389, W.DE 0.058389, peqDE 0.128720, Repeat
2015-05-23 12:33:22,201 Patch 58 of 64 DE 0.072995, W.DE 0.072995, peqDE 0.133085, Repeat (got worse)
2015-05-23 12:33:27,960 Patch 58 of 64 DE 0.072044, W.DE 0.072044, peqDE 0.139769, Repeat (got worse)
2015-05-23 12:33:33,719 Patch 58 of 64 DE 0.031834, W.DE 1.392589, peqDE 1.393452, Repeat
2015-05-23 12:33:39,477 Patch 58 of 64 DE 0.078606, W.DE 4.719636, peqDE 0.128720, Repeat (got worse)
2015-05-23 12:33:45,237 Patch 58 of 64 DE 0.105108, W.DE 0.105108, peqDE 0.168591, Repeat (got worse)
2015-05-23 12:33:50,995 Patch 58 of 64 DE 0.088734, W.DE 0.088734, peqDE 0.153434, Repeat
2015-05-23 12:33:56,755 Patch 58 of 64 DE 0.058389, W.DE 0.058389, W.peqDE 0.128720, Fail ( > 0.055188)
2015-05-23 12:34:02,792 Patch 59 of 64 DE 0.135693, W.DE 3.020544, peqDE 3.019560, Repeat
2015-05-23 12:34:08,553 Patch 59 of 64 DE 0.051407, W.DE 0.051407, peqDE 0.029742, Repeat
2015-05-23 12:34:32,577 Patch 59 of 64 DE 0.051407, W.DE 0.051407, peqDE 0.029742, Repeat
2015-05-23 12:34:38,767 Patch 59 of 64 DE 0.025714, W.DE 0.025714, W.peqDE 0.047912, OK ( < 0.050000)
2015-05-23 12:34:44,762 Patch 60 of 64 DE 0.066215, W.DE 21.472329, peqDE 21.472326, Repeat
2015-05-23 12:34:50,577 Patch 60 of 64 DE 0.103402, W.DE 6.949491, peqDE 6.949711, Repeat
2015-05-23 12:34:56,337 Patch 60 of 64 DE 0.084579, W.DE 4.803974, peqDE 4.804184, Repeat
2015-05-23 12:35:02,095 Patch 60 of 64 DE 0.069684, W.DE 0.069684, peqDE 0.079197, Repeat
2015-05-23 12:35:26,100 Patch 60 of 64 DE 0.069684, W.DE 0.069684, peqDE 0.079197, Repeat
2015-05-23 12:35:32,282 Patch 60 of 64 DE 0.071595, W.DE 8.176292, peqDE 0.079197, Repeat (got worse)
2015-05-23 12:35:38,041 Patch 60 of 64 DE 0.131936, W.DE 10.680631, peqDE 0.079197, Repeat (got worse)
2015-05-23 12:35:43,800 Patch 60 of 64 DE 0.051431, W.DE 9.656199, peqDE 0.079197, Repeat (got worse)
2015-05-23 12:35:49,558 Patch 60 of 64 DE 0.055389, W.DE 2.812526, peqDE 2.812861, Repeat
2015-05-23 12:35:55,317 Patch 60 of 64 DE 0.023392, W.DE 2.785667, peqDE 2.785722, Repeat
2015-05-23 12:36:01,076 Patch 60 of 64 DE 0.167168, W.DE 10.025102, peqDE 0.079197, Repeat (got worse)
2015-05-23 12:36:06,856 Patch 60 of 64 DE 0.069684, W.DE 0.069684, W.peqDE 0.079197, Fail ( > 0.050000)
2015-05-23 12:36:12,802 Patch 61 of 64 DE 0.099799, W.DE 15.283656, peqDE 15.283955, Repeat
2015-05-23 12:36:18,568 Patch 61 of 64 DE 0.059166, W.DE 5.723506, peqDE 5.723381, Repeat
2015-05-23 12:36:24,327 Patch 61 of 64 DE 0.044957, W.DE 0.044957, W.peqDE 0.041808, OK ( < 0.050000)
2015-05-23 12:36:30,250 Patch 62 of 64 DE 0.063937, W.DE 24.159951, peqDE 24.159924, Repeat
2015-05-23 12:36:36,011 Patch 62 of 64 DE 0.031306, W.DE 6.056478, peqDE 6.056444, Repeat
2015-05-23 12:37:00,014 Patch 62 of 64 DE 0.031306, W.DE 6.056478, peqDE 6.056444, Repeat
2015-05-23 12:37:06,203 Patch 62 of 64 DE 0.062630, W.DE 26.743523, peqDE 6.056444, Repeat (got worse)
2015-05-23 12:37:11,969 Patch 62 of 64 DE 0.063603, W.DE 7.065075, peqDE 7.065230, Repeat (got worse)
2015-05-23 12:37:17,736 Patch 62 of 64 DE 0.069990, W.DE 9.699714, peqDE 9.699708, Repeat (got worse)
2015-05-23 12:37:23,502 Patch 62 of 64 DE 0.033392, W.DE 2.758456, peqDE 2.758836, Repeat
2015-05-23 12:37:29,267 Patch 62 of 64 DE 0.045740, W.DE 0.045740, peqDE 0.068614, Repeat
2015-05-23 12:37:35,048 Patch 62 of 64 DE 0.054697, W.DE 0.054697, peqDE 0.076771, Repeat
2015-05-23 12:37:40,807 Patch 62 of 64 DE 0.046346, W.DE 25.595311, peqDE 0.068614, Repeat (got worse)
2015-05-23 12:37:46,568 Patch 62 of 64 DE 0.026241, W.DE 15.761124, peqDE 15.761195, Repeat
2015-05-23 12:37:52,332 Patch 62 of 64 DE 0.045740, W.DE 0.045740, W.peqDE 0.068614, Fail ( > 0.050000)
2015-05-23 12:37:58,177 Patch 63 of 64 DE 0.107010, W.DE 0.107010, peqDE 0.070682, Repeat
2015-05-23 12:38:22,173 Patch 63 of 64 DE 0.107010, W.DE 0.107010, peqDE 0.070682, Repeat
2015-05-23 12:38:28,355 Patch 63 of 64 DE 0.022374, W.DE 10.467839, peqDE 10.468054, Repeat
2015-05-23 12:38:34,121 Patch 63 of 64 DE 0.045021, W.DE 0.045021, peqDE 0.086982, Repeat
2015-05-23 12:38:39,917 Patch 63 of 64 DE 0.057709, W.DE 26.849092, peqDE 0.086982, Repeat (got worse)
2015-05-23 12:38:45,675 Patch 63 of 64 DE 0.036248, W.DE 29.163642, peqDE 0.086982, Repeat (got worse)
2015-05-23 12:38:51,434 Patch 63 of 64 DE 0.073884, W.DE 0.073884, peqDE 0.026817, Repeat (got worse)
2015-05-23 12:38:57,193 Patch 63 of 64 DE 0.093724, W.DE 81.953985, peqDE 0.086982, Repeat (got worse)
2015-05-23 12:39:02,975 Patch 63 of 64 DE 0.041675, W.DE 32.229442, peqDE 0.086982, Repeat (got worse)
2015-05-23 12:39:08,733 Patch 63 of 64 DE 0.039162, W.DE 32.766620, peqDE 0.086982, Repeat (got worse)
2015-05-23 12:39:14,492 Patch 63 of 64 DE 0.071839, W.DE 33.679686, peqDE 0.086982, Repeat (got worse)
2015-05-23 12:39:20,251 Patch 63 of 64 DE 0.045021, W.DE 0.045021, W.peqDE 0.086982, Fail ( > 0.050000)
2015-05-23 12:39:26,167 Patch 64 of 64 DE 0.007048, W.DE 0.007048, W.peqDE 0.007048, OK ( < 0.050000)
2015-05-23 12:39:26,168
2015-05-23 12:39:26,169 Brightness error = -0.033924 cd/m^2 (is 79.004316, should be 79.038239)
2015-05-23 12:39:26,170 White point error = 0.118328 deltaE
2015-05-23 12:39:26,171 Maximum neutral error (@ 0.502932) = 0.598883 deltaE
2015-05-23 12:39:26,171 Average neutral error = 0.283521 deltaE
2015-05-23 12:39:26,172 Failed to meet target 0.600000 delta E, got worst case 0.069684
2015-05-23 12:39:26,172 Number of measurements taken = 234
2015-05-23 12:39:26,173 Computing update to calibration curves…
2015-05-23 12:39:26,574 The instrument can be removed from the screen.
2015-05-23 16:31:50,160 dispcalGUI: Reached EOF (OK)
~~~~- This reply was modified on 2015-05-23 14:40:05 by *anonymous.
2015-05-23 at 18:49 #601I tried it again. Same result 🙁
2015-05-23 at 20:04 #602I couldn’t reproduce it here, so at the moment I’m not able to help I’m afraid. If you feel like it, you could do another run with diagnostic output (enable it in the “Options” menu, and set calibration speed to “Very high” to save time). It’ll likely fail as well, but the diagnostic output might contain information that allows to determine the cause of the failure.
2015-06-16 at 21:52 #603I have exactly the same system as the author and the same Colormunki photo is an identical issue, but I thought I have one error
2015-06-18 at 21:15 #604I also encountered the same problem but on a Mac Mini Late-2012 and with either the ColorMunki or i1Pro2.
I tried running in debug but at some point the window with the patch color disappear and the process goes on (or more likely the wait).
As of now I was not able to finish a calibration/profile. This machine is freshly re-installed and I’m using Argyll CMS 1.7.0.Anyway you should find attached the log with the debug information (for two different runs).
When the process is ‘stuck’ either dispcal/dispread was still running or nothing except the GUI. At one point I disconnected the device at it make no sense to measure the desktop background.Some screenshots when it’s waiting :
In case the attachment fails :
https://transfer.sh/BAzYf/dispcalgui-log.log
https://transfer.sh/ZlpLs/e2313hhttps://transfer.sh/%28/BAzYf/dispcalgui-log.log,/ZlpLs/e2313h%29.tar.gz
2015-06-18 at 21:53 #605Yours does not seem to be the same problem as reported by the OP.
I suspect USB problems. The more recent Mac Mini models have USB 3.0 ports. The ColorMunki on the other hand has been developed in the pre-USB 3 era.
When the process is ‘stuck’ either dispcal/dispread was still running or nothing except the GUI.
Are you sure about the latter? The progress dialog should always close automatically when the child process (dispcal/dispread) exits, and the only way I can see this not happening is if the system call to waitpid() still indicates that the child process is still alive while in reality it’s already dead.
From your first log (dispcalgui-log.log), it looks like the calibration ran through successfully, but profiling failed with an error due to the instrument not being found (I suspect an USB problem here).
The second log (e2313h) looks like the calibration finished just fine.
-