what verification settings to check to have reliable mesurement report

Home Forums Help and Support what verification settings to check to have reliable mesurement report

Viewing 2 posts - 1 through 2 (of 2 total)
  • Author
    Posts
  • #16942

    Keven
    Participant
    • Offline

    hi,

    I have calibrate my monitor and when finished it tell me i have a DeltaE*76 : average 0.14, max 1.48, RMS 0.30

    then just after that I do a measurement report and I get a result report that exceed nominal tolerance

    Average ΔE*00 =2.83

    Maximum ΔE*00=6.46

    I did try that 3 time to make sure I wasn’t messing anything with the instrument on the screen with the same result.

    then I tried to change some setting in the verification tab and I got a much better result

    Average ΔE*00=0.23

    Maximum ΔE*00=1.42

    the problem is i don’t know why the result change so much by changing some setting in the verification.

    ALSO i don’t know if the setting i use now are correct for my calibration, so i will link picture with all my setting to make it easier what setting is/was wrong and what is good.

    my goal is to have accurate color for an sRGB workflow. Most of my work is in lightroom and photoshop and publishing to the web, but i also do some video editing in premiere pro.

    Attachments:
    You must be logged in to view attached files.
    #16953

    Vincent
    Participant
    • Offline

    You are testing without GPU calibration. Uncheck simulation profile, it does not work as you think.