SpyderX – self-calibration fails

Home Forums Help and Support SpyderX – self-calibration fails

Viewing 2 posts - 16 through 17 (of 17 total)
  • Author
    Posts
  • #23941

    Florian Höch
    Administrator
    • Offline

    So, does this problem (behavior) seem to come from Argyll ?

    The offset is calculated by Argyll, and applied by it. The check for a (imo) reasonable threshold (15) is done in DisplayCAL, because (some) people kept ignoring the self-calibration message, leaving the instrument on screen, which then lead to totally bogus measurements 🙁

    As this offset is meant to counteract black level drift (rise) of the SpyderX, I was very surprised to see negative numbers there. I am unsure if this is a problem though – need to ask Graeme about it (by the way, if you want to do that, go ahead – I’m very short on time lately).

    And hey! Other poster with this problem… post a log so we can see what’s in there!

    Yes please. For this to be useful, it is actually necessary to enable Argyll diagnostic output, like you did.

    SpyderX Pro on Amazon  
    Disclosure: As an Amazon Associate I earn from qualifying purchases.

    #25029

    Nathan Sopeer
    Participant
    • Offline

    Hello, I just wanted to mention that I was having issues with this today and it is a little cool here, I just warmed up the spyderX for 10 seconds in the sun and worked all good after that.

Viewing 2 posts - 16 through 17 (of 17 total)