"Native" display parameters: how to use?

Home Forums Help and Support "Native" display parameters: how to use?

Viewing 2 posts - 1 through 2 (of 2 total)
  • Author
    Posts
  • #13737

    Alex Potemkin
    Participant
    • Offline

    Hello.

    I need to pass through characterisation(profiling) without calibration, preserve all settings lile white point, white level etc untouched.
    I another calibration software, I can use option “native” which mean that during the characterization this particular parameter will be used “as is” and no any correction will be applied via profile.

    But in DisplayCal I can’t see anything similar. The available options are only “As measured” or direct input of specific value.

    How to do it in ВС, Thank you!

    #13739

    Florian Höch
    Administrator
    • Offline

    “As measured” is what you want to use. Set all options on the calibration tab accordingly to skip 1D LUT calibration.

Viewing 2 posts - 1 through 2 (of 2 total)