My calibration gets wried. anyone know how to fix?
Attachments:
You must be
logged in to view attached files.
Make sure you disable any “dynamic” features of your display (dynamic dimming/contrast/backlight, energy saving, “Night mode” etc).
Are you use SpyderX pro?
it think you are same this problem with me.
My software use displaycal 3.8.5 ArgyllCMS 2.1.1 , My laptop is MacbookPro 15 this result in same laptop and same osx.
If i calibration with spyderxPro it’s same problem with this topic.
If i switch to spyderxPro with official software color back to normal.
If i calibration with i1-Display Pro everything fine color correct.
I try to measure after calibration with i1-DisplayPro and SpyderXPro this result in a attach file.
I try to change to calibration with windows10 but same result.
Attachments:
You must be
logged in to view attached files.
Yep, I call my friend yo use i1-Display and it’s work gorgeous.
But, My spyderX Pro get this result.
Are you doing the SpyderX black calibration correctly? You need to take the instrument off the screen when asked, attach the cap (it makes a clicking sound when it’s firmly shut), do the black calibration, then re-attach the SpyderX to the screen and continue.
OH!!!! It’s work!!!!! Thank you!!!!