i have Spyder3 probe and LCD Philips 246E7. After calibration report say ΔE 76 average 0.09, maximum 0.62 and RMS 0.18.
But when I do immediately verification, maximum ΔE is 11.38, and i dont have an idea where is problem.
In attachment is verification file, and screenshot with report after calibration.
Thanks for the advice
Attachments:
You must be
logged in to view attached files.
Hi,
you’re verifying against SpyderChecker colors, and some of them are seemingly out of gamut for your display.
Many thanks for answer, i did not think I could select another testchart and automatically select a spyderCHECKR because I have a spyder probe…..
The attachments includes additional measurements and looks very good.
And offcourse, really good software, i will have to donate some cash.
Attachments:
You must be
logged in to view attached files.