Maximum ΔE76 after calibration

Home Forums Help and Support Maximum ΔE76 after calibration

Viewing 3 posts - 1 through 3 (of 3 total)
  • Author
    Posts
  • #8129

    Fox25
    Participant
    • Offline

    Hello,

    i have Spyder3 probe and LCD Philips 246E7.  After calibration report say ΔE 76 average 0.09, maximum 0.62 and RMS 0.18.

    But when I do immediately verification, maximum ΔE is 11.38, and i dont have an idea where is problem.

    In attachment is verification file, and screenshot with report after calibration.

    Thanks for the advice

    Attachments:
    You must be logged in to view attached files.
    #8137

    Florian Höch
    Administrator
    • Offline

    Hi,

    you’re verifying against SpyderChecker colors, and some of them are seemingly out of gamut for your display.

    #8142

    Fox25
    Participant
    • Offline

    Many thanks for answer, i did not think I could select another testchart and automatically select a spyderCHECKR because I have a spyder probe…..

    The attachments includes additional measurements and looks very good.

    And offcourse, really good software, i will have to donate some cash.

    Attachments:
    You must be logged in to view attached files.
Viewing 3 posts - 1 through 3 (of 3 total)

You must be logged in to reply to this topic.