Instrument Access Failed Error

Home Forums Help and Support Instrument Access Failed Error

Viewing 2 posts - 1 through 2 (of 2 total)
  • Author
    Posts
  • #36678

    Jix
    Participant
    • Offline

    I recently bought SpyderX Elite, tried to calibrate my monitor (Lenovo ThinkVision P24q-10)  on my pc (Windows 10) and everything worked alright with spyder’s own calibration software. When I downloaded displaycal, I manually changed driver on my SpyderX to Argyll_V2.3.1 with disabled driver signature enforcement. When I start the callibration the error appears and I cant really do anything. It only working normally when I unplug my spyderx from usb and plug it in again and I manage to start the calibration and also get to the patch checking, but not for long, because same error appears after that again. I really dont know what to do. Please help!

    I attached my log file below.

    Attachments:
    You must be logged in to view attached files.

    SpyderX Pro on Amazon  
    Disclosure: As an Amazon Associate I earn from qualifying purchases.

    #36705

    Jix
    Participant
    • Offline

    SOLVED IT! So the problem was actually a bug of Argyll V2.3.1 . I tried to change 2.3.1 driver to 2.3.0 and everything worked flawlessly!

Viewing 2 posts - 1 through 2 (of 2 total)

You must be logged in to reply to this topic.

Log in or Register

Display Calibration and Characterization powered by ArgyllCMS