I have this error when I want to create a colorimeter correction with my i1DisplayPro OEM & i1Pro 2 Spectrophotometer Rev E. (I have installed Installing ArgyllCMS instrument drivers). When I click on “Measure” with i1Pro2 say “Error – New_disprd failed with” Instrument access failed”.
What am I doing wrong? I can not find an option to initialize the spectrophotometer! Thks in advance!
1) Tools -> Instrument -> Spectrometer self-calibration
2) Tools -> Correction -> Create colorimeter correction -> Mesasure
I activate Argyll CMS diagnostic output. I attached the result file DisplayCAL.log.
Attachments:
You must be
logged in to view attached files.
init_inst returned ‘Hardware Failure’ (Instrument calibration info is missing or corrupted)
Sounds bad. If you switch back to the X-Rite driver and run X-Rite i1 Diagnostic, does it pass?
Do measurements work in i1 Profiler?
Does using a different measurement mode help? (i.e. use the default “LCD (generic) adaptive”)
Do measurements work in i1 Profiler?
Does using a different measurement mode help? (i.e. use the default “LCD (generic) adaptive”)
I do not have that mode, I have “Direct view mode”.
Do measurements work in i1 Profiler?
Does using a different measurement mode help? (i.e. use the default “LCD (generic) adaptive”)
That seems to be CalMAN?
Yes, those are the modes that appear in Calman (as an example) and these are the modes that appear in DisplayCal… I try with all the modes that appear in DisplayCal, without lucky, same error.
X-Rite i1 Diagnostic log file
Attachments:
You must be
logged in to view attached files.