Error message at the beginning

Home Forums Help and Support Error message at the beginning

Viewing 2 posts - 1 through 2 (of 2 total)
  • Author
    Posts
  • #13762

    Heldwin
    Participant
    • Offline

    Hello,

    I changed my computer and I wanted to redo my screen calibration.

    My probe is an i1 display Pro. I have this error message: “New_disprd failed with ‘instrument access failed‘” and I can not solve the problem.
    I have tried several USB ports (they are all USB 3.0).
    Here is the log following the error message:

    14: 48: 26,848 Calibrate and characterize
    14: 48: 26.933
    14: 48: 27,002 Detecting output levels range ...
    14: 48: 27,006 --------------------------------------------- -----------------------------------
    14: 48: 27,006 Log Session: 0_16
    14: 48: 27.006
    14: 48: 27,006 Working Directory:
    14: 48: 27,006 c: \
    14: 48: 27,006 users \
    14: 48: 27,006 heldwin \
    14: 48: 27,007 appdata \
    14: 48: 27,007 local \
    14: 48: 27,009 temp \
    14: 48: 27,009 DisplayCAL-7ab5ny \
    14: 48: 27.009
    14: 48: 27,009 Command line:
    14: 48: 27,009 E: \ Argyll_V2.0.1 \ bin \ dispread.exe
    14: 48: 27,010 -v
    14: 48: 27,010 -k
    14: 48: 27,010 "C: \ Program Files (x86) \ DisplayCAL \ linear.cal"
    14: 48: 27,010 -d1
    14: 48: 27,012 -c1
    14: 48: 27,012 -yn
    14: 48: 27.012 "-P0.497918834547.0.495012468828,1.48717948718"
    14: 48: 27,013 0_16
    14: 48: 27.013
    14: 48: 27,263 DisplayCAL: Starting interaction with subprocess
    14: 48: 27,266 Number of patches = 3
    14: 48: 27,266 Setting up the instrument
    14: 48: 27,375 hid_open_port: Failed to open path ↲
                 ↳ '\\? \ Hid # vid_0765 & pid_5020 # 6 & 1886a4ad & 0 & 0000 # {4d1e55b2-f16f-11cf- ↲
                 ↳ 88cb-001111000030} 'with err 24
    14: 48: 27,375 hid_open_port: Failed to open path ↲
                 ↳ '\\? \ Hid # vid_0765 & pid_5020 # 6 & 1886a4ad & 0 & 0000 # {4d1e55b2-f16f-11cf- ↲
                 ↳ 88cb-001111000030} 'with err 24
    14: 48: 27,375 dispread: Error - new_disprd failed with 'Instrument Access Failed'
    14: 48: 27.375
    14: 48: 27,480 DisplayCAL: Reached EOF (OK)
    14: 48: 27,480 ... interrupted.

    Do you have a solution to solve the problem?
    Thank you

    Calibrite Display Pro HL on Amazon  
    Disclosure: As an Amazon Associate I earn from qualifying purchases.

    #13800

    Florian Höch
    Administrator
    • Offline

    Hi,

    disable X-Rite devices service if it is installed, and make sure no other calibration software is running.

Viewing 2 posts - 1 through 2 (of 2 total)