Home › Forums › Help and Support › Why use the K-10 probe, the program will terminate it?
- This topic has 4 replies, 2 voices, and was last updated 8 years, 9 months ago by
Florian Höch.
-
AuthorPosts
-
2017-08-23 at 2:31 #8519
what happen ?
I use K-10“2017-08-22 09:22:33,276 ================================================== ==============================
2017-08-22 09:22:33,276 DisplayCAL.exe 3.3.3.0 2017-08-08T18:45:19.543Z
2017-08-22 09:22:33,276 Windows 10 Home China x86_64
2017-08-22 09:22:33,276 Python 2.7.12 (v2.7.12:d33e0cf91556, Jun 27 2016, 15:19:22) [MSC v.1500 32 bit (Intel)]
2017-08-22 09:22:33,276 Faulthandler 2.6
2017-08-22 09:22:33,510 Imported existing <module ‘comtypes.gen’ from ‘C:\Program Files (x86)\DisplayCAL\lib\library.zip\comtypes\gen\__in it__.pyc’>
2017-08-22 09:22:33,510 Using writeable comtypes cache directory: ‘c:\users\qiuji\appdata\local\temp\comtypes_cache\ DisplayCAL-27’
2017-08-22 09:22:33,510 wxPython 3.0.2.0 msw (classic)
2017-08-22 09:22:33,510 Encoding: cp936
2017-08-22 09:22:33,510 File system encoding: mbcs
2017-08-22 09:22:33,698 Starting up…
2017-08-22 09:22:33,855 Audio module: pyglet 1.2.2
2017-08-22 09:22:35,543 Initializing GUI…
2017-08-22 09:22:35,933
2017-08-22 09:22:37,619 …ok.
2017-08-22 09:22:37,619 Ready.
2017-08-22 09:22:37,885 Setting up scripting host at 127.0.0.1:3866
2017-08-22 09:22:37,994 Check for application update…
2017-08-22 09:22:38,813 DisplayCAL is up-to-date.
2017-08-22 09:22:58,329 Enumerating display devices and communication ports…
2017-08-22 09:22:58,331 C:\Argyll_V1.9.2\bin
2017-08-22 09:22:58,331 DisplayCAL-apply-profiles setcfg argyll.dir “C:\Argyll_V1.9.2\bin” force returned ok
2017-08-22 09:23:03,263 ArgyllCMS 1.9.2
2017-08-22 09:23:03,266 …ok.
2017-08-22 09:23:03,381 Checking video card gamma table access for display 1…
2017-08-22 09:23:03,522 Verify: ‘test.cal’ IS loaded (discrepancy 0.0%)
2017-08-22 09:23:03,569 …ok.
2017-08-22 09:23:03,571 Checking video card gamma table access for display 2…
2017-08-22 09:23:03,743 Verify: ‘test.cal’ IS loaded (discrepancy 0.0%)
2017-08-22 09:23:03,822 …ok.
2017-08-22 09:23:07,197 A change in the display device configuration has been detected.
2017-08-22 09:23:07,263 A change in the instrument/port configuration has been detected.
2017-08-22 09:23:40,532 ——————————————————————————–
2017-08-22 09:23:40,533 Calibrate & profile
2017-08-22 09:23:41,575 Detecting output levels range…
2017-08-22 09:23:41,579 ——————————————————————————–
2017-08-22 09:23:41,581 Session log: 0_16
2017-08-22 09:23:41,581
2017-08-22 09:23:41,582 Working directory:
2017-08-22 09:23:41,582 c:\
2017-08-22 09:23:41,584 users\
2017-08-22 09:23:41,585 qiuji\
2017-08-22 09:23:41,585 appdata\
2017-08-22 09:23:41,585 local\
2017-08-22 09:23:41,585 temp\
2017-08-22 09:23:41,586 DisplayCAL-hpsaxn\
2017-08-22 09:23:41,586
2017-08-22 09:23:41,588 Command line:
2017-08-22 09:23:41,588 dispread.exe
2017-08-22 09:23:41,588 -v
2017-08-22 09:23:41,588 -k
2017-08-22 09:23:41,589 “C:\Program Files (x86)\DisplayCAL\linear.cal”
2017-08-22 09:23:41,589 -d1
2017-08-22 09:23:41,591 -c1
2017-08-22 09:23:41,592 -yF
2017-08-22 09:23:41,592 “-P0.500165837479,0.500374531835,1.5”
2017-08-22 09:23:41,594 0_16
2017-08-22 09:23:41,594
2017-08-22 09:23:41,926 DisplayCAL: Starting interaction with subprocess
2017-08-22 09:23:43,954 Number of patches = 3
2017-08-22 09:23:43,957 Setting up the instrument
2017-08-22 09:23:45,811 dispread: Error – new_disprd failed with ‘Instrument Access Failed’
2017-08-22 09:23:45,812
2017-08-22 09:23:45,914 DisplayCAL: Reached EOF (OK)
2017-08-22 09:23:45,917 …aborted.
2017-08-22 09:23:57,494 Executing BaseApp.OnExit()
2017-08-22 09:23:57,494 Running application exit handlers
2017-08-22 09:23:57,496 Waiting for thread ScriptingHost.ConnectionHandler to exit
2017-08-22 09:23:57,581 ScriptingHost.ConnectionHandler exited
2017-08-22 09:23:57,582 Exiting DisplayCAL
2017-08-22 09:23:57,825 Ran application exit handlers”Attachments:
You must be logged in to view attached files.2017-08-24 at 16:17 #8554Make sure nothing else is accessing the instrument. Did you install the FTDI virtual com port driver that came with the K-10? Is it visible in Device Manager?
2017-08-25 at 3:40 #8572the driver is ok ,I can USE K-10 in calman,lightspace .
2017-08-25 at 3:47 #8573C:\Argyll_V1.9.2\bin>spotread -D8
Argyll ‘V1.9.2’ Build ‘MSWin 64 bit’ System ‘Windows V10.0 SP 0’
icoms_refresh_paths: called with mask = 131071
icoms_refresh_paths: looking for HID device
hid_get_paths: called
found HID device ‘\\?\hid#vid_046d&pid_c31c&mi_00#7&38bf1b23&0&0000#{4d1e55b2-f16f-11cf-88cb-001111000030}’, inst 1 but not one we’re looking for
found HID device ‘\\?\hid#vid_046d&pid_c31c&mi_01&col01#7&258efee5&0&0000#{4d1e55b2-f16f-11cf-88cb-001111000030}’, inst 2 but not one we’re looking for
found HID device ‘\\?\hid#vid_046d&pid_c31c&mi_01&col02#7&258efee5&0&0001#{4d1e55b2-f16f-11cf-88cb-001111000030}’, inst 3 but not one we’re looking for
found HID device ‘\\?\hid#vid_046d&pid_c31c&mi_01&col03#7&258efee5&0&0002#{4d1e55b2-f16f-11cf-88cb-001111000030}’, inst 4 but not one we’re looking for
found HID device ‘\\?\hid#vid_093a&pid_2510#6&201d83a&0&0000#{4d1e55b2-f16f-11cf-88cb-001111000030}’, inst 5 but not one we’re looking for
icoms_get_paths: returning 0 paths and ICOM_OK
icoms_refresh_paths: looking for USB device
usb_get_paths opening device ‘\\.\libusb0-0001’
usb_get_paths opening device ‘\\.\libusb0-0002’
usb_get_paths opening device ‘\\.\libusb0-0003’
usb_get_paths opening device ‘\\.\libusb0-0004’
usb_get_paths opening device ‘\\.\libusb0-0005’
usb_get_paths opening device ‘\\.\libusb0-0006’
usb_get_paths opening device ‘\\.\libusb0-0007’
usb_get_paths opening device ‘\\.\libusb0-0008’
usb_get_paths opening device ‘\\.\libusb0-0009’
usb_get_paths opening device ‘\\.\libusb0-0010’
usb_get_paths opening device ‘\\.\libusb0-0011’
usb_get_paths opening device ‘\\.\libusb0-0012’
usb_get_paths opening device ‘\\.\libusb0-0013’
usb_get_paths opening device ‘\\.\libusb0-0014’
usb_get_paths opening device ‘\\.\libusb0-0015’
usb_get_paths opening device ‘\\.\libusb0-0016’
icoms_refresh_paths: now got 0 paths
icoms_refresh_paths: looking for serial ports
serial_get_paths: called with mask = 131071
serial_get_paths: looking up the registry for serial ports
serial_get_paths: looking through all the values in the SERIALCOMM key
serial_get_paths: checking ‘VCP0’
icompaths_add_serial got ‘COM4’ dctype 0x100
icompaths_add_serial returning ‘COM4’ dctype 0xf0501
serial_get_paths: Added ‘VCP0’ path ‘COM4’ dctype 0x100
new_icoms ‘COM4’ itype ‘Unknown’ dctype 0xf0501
icom_copy_path_to_icom ‘COM4’ returning dctype 0xf0501
fast_ser_inst_type: on ‘COM4’ dctype 0xf0501
fser_inst_type: Trying different baud rates (2000 msec to go) Path COM4
Trying 9600 baud
icoms_set_ser_port: About to set port characteristics:
Port name = COM4
Flow control = 1
Baud Rate = 9600
Parity = 1
Stop bits = 1
Word length = 4
Open delay = 0 ms
icoms_set_ser_port: Make sure serial port is open
icoms_set_ser_port: about to open serial port ‘COM4’
icoms_set_ser_port: port characteristics set ok
icoms_write_read: calledicoms_ser_read: bytes 500, ttop 20, ntc 500
icoms_ser_read: read 0 total bytes with 2 reads
icoms_ser_read: took 32 msec
icoms_ser_read: timeout, took 32 msec out of 20
icoms_ser_read: took 32 msec, returning ” ICOM err 0x4000icoms_ser_write: writing ‘;D024^M^J’
icoms_ser_write: ep 0x0, bytes 7, ttop 200, quant 0
icoms_ser_write: wrote 7 bytes
icoms_ser_write: took 0 msec, returning ICOM err 0x0icoms_ser_read: bytes 2057, ttop 200, ntc 1
icoms_ser_read: read 7 bytes, rbuf = ‘;D024^M^J’
icoms_ser_read: tc count 1
icoms_ser_read: read 7 total bytes with 1 reads
icoms_ser_read: took 31 msec
icoms_ser_read: took 31 msec, returning ‘;D024^M^J’ ICOM err 0x0
icoms_write_read: returning 0x0
fser_inst_type: got 7 bytes
fser_inst_type: Not Klein response
Trying 921600 baud
icoms_set_ser_port: About to set port characteristics:
Port name = COM4
Flow control = 1
Baud Rate = 921600
Parity = 1
Stop bits = 1
Word length = 4
Open delay = 0 ms
icoms_set_ser_port: closing port ‘COM4’
icoms_set_ser_port: Make sure serial port is open
icoms_set_ser_port: about to open serial port ‘COM4’
icoms_set_ser_port: port characteristics set ok
icoms_write_read: calledicoms_ser_read: bytes 500, ttop 20, ntc 500
icoms_ser_read: read 0 total bytes with 1 reads
icoms_ser_read: took 31 msec
icoms_ser_read: timeout, took 31 msec out of 20
icoms_ser_read: took 31 msec, returning ” ICOM err 0x4000icoms_ser_write: writing ‘*idn?^M’
icoms_ser_write: ep 0x0, bytes 6, ttop 100, quant 0
icoms_ser_write: wrote 6 bytes
icoms_ser_write: took 0 msec, returning ICOM err 0x0icoms_ser_read: bytes 2058, ttop 100, ntc 1
icoms_ser_read: read 6 bytes, rbuf = ‘*idn?^M’
icoms_ser_read: tc count 1
icoms_ser_read: read 6 total bytes with 1 reads
icoms_ser_read: took 31 msec
icoms_ser_read: took 31 msec, returning ‘*idn?^M’ ICOM err 0x0
icoms_write_read: returning 0x0
Trying 115200 baud
icoms_set_ser_port: About to set port characteristics:
Port name = COM4
Flow control = 1
Baud Rate = 115200
Parity = 1
Stop bits = 1
Word length = 4
Open delay = 0 ms
icoms_set_ser_port: closing port ‘COM4’
icoms_set_ser_port: Make sure serial port is open
icoms_set_ser_port: about to open serial port ‘COM4’
icoms_set_ser_port: port characteristics set ok
icoms_write_read: calledicoms_ser_read: bytes 500, ttop 20, ntc 500
icoms_ser_read: read 0 total bytes with 2 reads
icoms_ser_read: took 31 msec
icoms_ser_read: timeout, took 31 msec out of 20
icoms_ser_read: took 31 msec, returning ” ICOM err 0x4000icoms_ser_write: writing ‘*idn?^M’
icoms_ser_write: ep 0x0, bytes 6, ttop 100, quant 0
icoms_ser_write: wrote 6 bytes
icoms_ser_write: took 0 msec, returning ICOM err 0x0icoms_ser_read: bytes 2058, ttop 100, ntc 1
icoms_ser_read: read 6 bytes, rbuf = ‘*idn?^M’
icoms_ser_read: tc count 1
icoms_ser_read: read 6 total bytes with 1 reads
icoms_ser_read: took 32 msec
icoms_ser_read: took 32 msec, returning ‘*idn?^M’ ICOM err 0x0
icoms_write_read: returning 0x0
Trying 38400 baud
icoms_set_ser_port: About to set port characteristics:
Port name = COM4
Flow control = 1
Baud Rate = 38400
Parity = 1
Stop bits = 1
Word length = 4
Open delay = 0 ms
icoms_set_ser_port: closing port ‘COM4’
icoms_set_ser_port: Make sure serial port is open
icoms_set_ser_port: about to open serial port ‘COM4’
icoms_set_ser_port: port characteristics set ok
icoms_write_read: called
icoms_ser_read: bytes 500, ttop 20, ntc 500
icoms_ser_read: read 0 total bytes with 2 reads
icoms_ser_read: took 31 msec
icoms_ser_read: timeout, took 31 msec out of 20
icoms_ser_read: took 31 msec, returning ” ICOM err 0x4000icoms_ser_write: writing ‘7e 00 02 00’
icoms_ser_write: ep 0x0, bytes 4, ttop 100, quant 0
icoms_ser_write: wrote 4 bytes
icoms_ser_write: took 0 msec, returning ICOM err 0x0icoms_ser_read: bytes 2058, ttop 100, ntc 4
icoms_ser_read: read 4 bytes, rbuf = ‘~’
icoms_ser_read: read 4 total bytes with 1 reads
icoms_ser_read: took 31 msec
icoms_ser_read: took 31 msec, returning ‘7e 00 02 00’ ICOM err 0x0
icoms_write_read: returning 0x0
icoms_write_read: calledicoms_ser_read: bytes 500, ttop 20, ntc 500
icoms_ser_read: read 0 total bytes with 2 reads
icoms_ser_read: took 31 msec
icoms_ser_read: timeout, took 31 msec out of 20
icoms_ser_read: took 31 msec, returning ” ICOM err 0x4000icoms_ser_write: writing ‘*idn?^M’
icoms_ser_write: ep 0x0, bytes 6, ttop 100, quant 0
icoms_ser_write: wrote 6 bytes
icoms_ser_write: took 0 msec, returning ICOM err 0x0icoms_ser_read: bytes 2058, ttop 100, ntc 1
icoms_ser_read: read 6 bytes, rbuf = ‘*idn?^M’
icoms_ser_read: tc count 1
icoms_ser_read: read 6 total bytes with 1 reads
icoms_ser_read: took 32 msec
icoms_ser_read: took 32 msec, returning ‘*idn?^M’ ICOM err 0x0
icoms_write_read: returning 0x0
fser_inst_type: Instrument type is ‘Unknown Instrument’
icoms_del: called
icoms_del: closing port
serial_get_paths: Identified ‘Unknown’ dctype 0xf0501
serial_get_paths: checking ‘VCP4’
icompaths_add_serial got ‘COM3’ dctype 0x100
icompaths_add_serial returning ‘COM3’ dctype 0xf0501
serial_get_paths: Added ‘VCP4’ path ‘COM3’ dctype 0x100
new_icoms ‘COM3’ itype ‘Unknown’ dctype 0xf0501
icom_copy_path_to_icom ‘COM3’ returning dctype 0xf0501
fast_ser_inst_type: on ‘COM3’ dctype 0xf0501
fser_inst_type: Trying different baud rates (2000 msec to go) Path COM3
Trying 9600 baud
icoms_set_ser_port: About to set port characteristics:
Port name = COM3
Flow control = 1
Baud Rate = 9600
Parity = 1
Stop bits = 1
Word length = 4
Open delay = 0 ms
icoms_set_ser_port: Make sure serial port is open
icoms_set_ser_port: about to open serial port ‘COM3’
icoms_set_ser_port: port characteristics set ok
icoms_write_read: calledicoms_ser_read: bytes 500, ttop 20, ntc 500
icoms_ser_read: read 0 total bytes with 2 reads
icoms_ser_read: took 31 msec
icoms_ser_read: timeout, took 31 msec out of 20
icoms_ser_read: took 31 msec, returning ” ICOM err 0x4000icoms_ser_write: writing ‘;D024^M^J’
icoms_ser_write: ep 0x0, bytes 7, ttop 200, quant 0
icoms_ser_write: wrote 7 bytes
icoms_ser_write: took 0 msec, returning ICOM err 0x0icoms_ser_read: bytes 2057, ttop 200, ntc 1
icoms_ser_read: read 174 bytes, rbuf = ‘D4CP2000 //’
icoms_ser_read: tc count 0
icoms_ser_read: read 174 total bytes with 1 reads
icoms_ser_read: took 203 msec
icoms_ser_read: timeout, took 203 msec out of 200
icoms_ser_read: took 203 msec, returning ‘D4CP2000 //’ ICOM err 0x4000
icoms_write_read: returning 0x4000
fser_inst_type: got 24 bytes
fser_inst_type: Looks like it may be a Kleinicoms_ser_read: bytes 2058, ttop 100, ntc 2058
icoms_ser_read: read 123 bytes, rbuf = ”
icoms_ser_read: read 92 bytes, rbuf = ”
icoms_ser_read: read 215 total bytes with 2 reads
icoms_ser_read: took 204 msec
icoms_ser_read: timeout, took 204 msec out of 100
icoms_ser_read: took 204 msec, returning ’00 00 00 00 00 00 00 00 00 00 00 00 00 00 00 5f 0b 07 64 00 07 6c e4 07 28 f6 fa 50 00 03 45 e3 03 c2 60 02 aa d1 01 d1 f7 01 54 86 02 2b 65 fd a3 63 fe a9 e8 00 6c 9d 01 49 7d 01 2e 3e fe 24 …’ ICOM err 0x4000icoms_ser_read: bytes 2058, ttop 100, ntc 2058
icoms_ser_read: read 123 bytes, rbuf = ‘0821-tw8300 //’
icoms_ser_read: read 123 total bytes with 1 reads
icoms_ser_read: took 109 msec
icoms_ser_read: timeout, took 109 msec out of 100
icoms_ser_read: took 109 msec, returning ’30 38 32 31 2d 74 77 38 33 30 30 20 20 20 20 20 20 2f 2f 00 00 00 00 00 00 00 00 00 00 00 00 00 00 00 00 00 00 00 00 00 00 00 00 00 00 00 00 00 00 00 00 00 00 00 00 00 5f 0b 07 64 00 07 6c e4 …’ ICOM err 0x4000icoms_ser_read: bytes 2058, ttop 100, ntc 2058
icoms_ser_read: read 107 bytes, rbuf = ‘2^A\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377’
icoms_ser_read: read 92 bytes, rbuf = ‘2^A\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377′
icoms_ser_read: read 199 total bytes with 2 reads
icoms_ser_read: took 203 msec
icoms_ser_read: timeout, took 203 msec out of 100
icoms_ser_read: took 203 msec, returning ’32 01 ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff …’ ICOM err 0x4000icoms_ser_read: bytes 2058, ttop 100, ntc 2058
icoms_ser_read: read 108 bytes, rbuf = ‘\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377’
icoms_ser_read: read 108 total bytes with 1 reads
icoms_ser_read: took 109 msec
icoms_ser_read: timeout, took 109 msec out of 100
icoms_ser_read: took 109 msec, returning ‘ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff …’ ICOM err 0x4000icoms_ser_read: bytes 2058, ttop 100, ntc 2058
icoms_ser_read: read 92 bytes, rbuf = ‘\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377’
icoms_ser_read: read 107 bytes, rbuf = ‘\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377’
icoms_ser_read: read 199 total bytes with 2 reads
icoms_ser_read: took 203 msec
icoms_ser_read: timeout, took 203 msec out of 100
icoms_ser_read: took 203 msec, returning ‘ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff …’ ICOM err 0x4000icoms_ser_read: bytes 2058, ttop 100, ntc 2058
icoms_ser_read: read 108 bytes, rbuf = ‘\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377’
icoms_ser_read: read 92 bytes, rbuf = ‘\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377’
icoms_ser_read: read 200 total bytes with 2 reads
icoms_ser_read: took 203 msec
icoms_ser_read: timeout, took 203 msec out of 100
icoms_ser_read: took 203 msec, returning ‘ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff …’ ICOM err 0x4000icoms_ser_read: bytes 2058, ttop 100, ntc 2058
icoms_ser_read: read 67 bytes, rbuf = ‘\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377\377<0>’
icoms_ser_read: read 67 total bytes with 2 reads
icoms_ser_read: took 187 msec
icoms_ser_read: timeout, took 187 msec out of 100
icoms_ser_read: took 187 msec, returning ‘ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff ff …’ ICOM err 0x4000
icoms_ser_read: bytes 2058, ttop 100, ntc 2058
icoms_ser_read: read 0 total bytes with 1 reads
icoms_ser_read: took 110 msec
icoms_ser_read: timeout, took 110 msec out of 100
icoms_ser_read: took 110 msec, returning ” ICOM err 0x4000
icoms_write_read: calledicoms_ser_read: bytes 500, ttop 20, ntc 500
icoms_ser_read: read 0 total bytes with 2 reads
icoms_ser_read: took 47 msec
icoms_ser_read: timeout, took 47 msec out of 20
icoms_ser_read: took 47 msec, returning ” ICOM err 0x4000icoms_ser_write: writing ‘P0^M’
icoms_ser_write: ep 0x0, bytes 3, ttop 100, quant 0
icoms_ser_write: wrote 3 bytes
icoms_ser_write: took 0 msec, returning ICOM err 0x0icoms_ser_read: bytes 2058, ttop 100, ntc 1
icoms_ser_read: read 21 bytes, rbuf = ‘P0K-10 C001114 <0>’
icoms_ser_read: tc count 1
icoms_ser_read: read 21 total bytes with 1 reads
icoms_ser_read: took 47 msec
icoms_ser_read: took 47 msec, returning ‘P0K-10 C001114 <0>’ ICOM err 0x0
icoms_write_read: returning 0x0
fser_inst_type: found Klein K1/K8/K10
fser_inst_type: Instrument type is ‘Klein K-10’
icompaths_set_serial_itype ‘COM3 (Klein K-10)’ returning dctype 0x10501
icoms_del: called
icoms_del: closing port
serial_get_paths: Identified ‘K-10’ dctype 0x10501
serial_get_paths: got ERROR_NO_MORE_ITEMS
icoms_refresh_paths: we now have 2 devices and are about to sort them
icompaths_make_dslists ‘COM3 (Klein K-10)’ dctype 0x10501
icompaths_make_dslists ‘COM4’ dctype 0xf0501
icoms_refresh_paths: returning 2 paths and ICOM_OK
new_inst: called with path ‘COM3 (Klein K-10)’ type ‘K-10’
new_icoms ‘COM3 (Klein K-10)’ itype ‘K-10’ dctype 0x10501
icom_copy_path_to_icom ‘COM3 (Klein K-10)’ returning dctype 0x10501
k10_init_coms: About to init Serial I/O
k10_init_coms: Trying different baud rates (500 msec to go)
k10_init_coms: Trying 9600 baud, 500 msec to go
icoms_set_ser_port: About to set port characteristics:
Port name = COM3 (Klein K-10)
Flow control = 4
Baud Rate = 9600
Parity = 1
Stop bits = 1
Word length = 4
Open delay = 0 ms
icoms_set_ser_port: Make sure serial port is open
icoms_set_ser_port: about to open serial port ‘COM3’
icoms_set_ser_port: port characteristics set ok
icoms_write_read: calledicoms_ser_write: writing ‘P0^M’
icoms_ser_write: ep 0x0, bytes 3, ttop 500, quant 0
icoms_ser_write: wrote 3 bytes
icoms_ser_write: took 0 msec, returning ICOM err 0x0icoms_ser_read: bytes 500, ttop 500, ntc 21
icoms_ser_read: read 0 total bytes with 1 reads
icoms_ser_read: took 500 msec
icoms_ser_read: timeout, took 500 msec out of 500
icoms_ser_read: took 500 msec, returning ” ICOM err 0x4000
icoms_write_read: returning 0x4000
error code 0xff02
k10_init_coms: failed to establish coms
k10_init_coms: About to init Serial I/O
k10_init_coms: Trying different baud rates (500 msec to go)
k10_init_coms: Trying 9600 baud, 484 msec to go
icoms_set_ser_port: About to set port characteristics:
Port name = COM3 (Klein K-10)
Flow control = 4
Baud Rate = 9600
Parity = 1
Stop bits = 1
Word length = 4
Open delay = 0 ms
icoms_set_ser_port: closing port ‘COM3 (Klein K-10)’
icoms_set_ser_port: Make sure serial port is open
icoms_set_ser_port: about to open serial port ‘COM3’
icoms_set_ser_port: port characteristics set ok
icoms_write_read: calledicoms_ser_write: writing ‘P0^M’
icoms_ser_write: ep 0x0, bytes 3, ttop 500, quant 0
icoms_ser_write: wrote 3 bytes
icoms_ser_write: took 0 msec, returning ICOM err 0x0icoms_ser_read: bytes 500, ttop 500, ntc 21
icoms_ser_read: read 0 total bytes with 1 reads
icoms_ser_read: took 500 msec
icoms_ser_read: timeout, took 500 msec out of 500
icoms_ser_read: took 500 msec, returning ” ICOM err 0x4000
icoms_write_read: returning 0x4000
error code 0xff02
k10_init_coms: failed to establish coms
Failed to initialise communications with instrument
or wrong instrument or bad configuration!
(‘Communications failure’ + ‘No device error’)
icoms_del: called
icoms_del: closing port2017-08-30 at 0:24 #8622Please contact Graeme Gill, the author of ArgyllCMS, and send him the above debug log.
-