Error – new_disprd() failed with 'Instrument Access Failed'

Home Forums Help and Support Error – new_disprd() failed with 'Instrument Access Failed'

Viewing 2 posts - 1 through 2 (of 2 total)
  • Author
    Posts
  • #3469

    Luke
    Participant
    • Offline

    Hello!

    I’ve been struggling for hours trying to get DisplayCAL to function properly. I am using it with a Colormonki Display. I install that and Argyll (via the instructions on their website) and I keep getting this same error over and over, no matter what I try. It works just fine in X-Rite. I have it in a USB 2.0. I restarted, reinstalled, et al to no avail. Any help is much appreciated. My log is below. Thanks!

    21:54:01,338 Calibrate & profile
    21:54:02,148 ——————————————————————————–
    21:54:02,148 Session log: XB271HU #1 2016-06-29 21-53 120cdm² D6500 2.2 F-S XYZLUT+MTX.log
    21:54:02,148
    21:54:02,148 Working directory:
    21:54:02,151 c:\
    21:54:02,151 users\
    21:54:02,151 luke\
    21:54:02,151 appdata\
    21:54:02,153 local\
    21:54:02,153 temp\
    21:54:02,153 DisplayCAL-et83fb\
    21:54:02,154
    21:54:02,154 Command line:
    21:54:02,154 dispcal.exe
    21:54:02,155 -v2
    21:54:02,155 -d1
    21:54:02,155 -c1
    21:54:02,157 -yn
    21:54:02,157 “-P0.413897280967,0.521495327103,2.10955710956”
    21:54:02,157 -X
    21:54:02,157 WGCCFLFamily_07Feb11.ccss
    21:54:02,157 -ql
    21:54:02,157 -t6500
    21:54:02,158 -b120.0
    21:54:02,158 -g2.2
    21:54:02,158 -f1.0
    21:54:02,158 -k0
    21:54:02,160 -A4.0
    21:54:02,160 -b
    21:54:02,160 “XB271HU #1 2016-06-29 21-53 120cdm² D6500 2.2 F-S XYZLUT+MTX”
    21:54:02,161
    21:54:02,456 DisplayCAL: Starting interaction with subprocess
    21:54:02,459 Argyll ‘V1.8.3’ Build ‘MSWin 64 bit’ System ‘Windows V10.0 SP 0’
    21:54:02,459 Setting up the instrument
    21:54:02,539 hid_open_port: Failed to open path
    21:54:02,539 ‘\\?\hid#vid_0765&pid_5020#6&2be69897&0&0000#{4d1e55b2-f16f-11cf-88cb-001111000030}’
    21:54:02,539 30}’
    21:54:02,539 with err 24
    21:54:02,539 hid_open_port: Failed to open path
    21:54:02,539 ‘\\?\hid#vid_0765&pid_5020#6&2be69897&0&0000#{4d1e55b2-f16f-11cf-88cb-001111000030}’
    21:54:02,539 30}’
    21:54:02,539 with err 24
    21:54:02,540 dispcal: Error – new_disprd() failed with ‘Instrument Access Failed’
    21:54:02,540
    21:54:02,644 DisplayCAL: Reached EOF (OK)
    21:54:02,647 …aborted.

    #3475

    Florian Höch
    Administrator
    • Offline

    It works just fine in X-Rite

    Stop and disable the “X-Rite Device Services Manager”, and make sure no other software is accessing the device (i.e. close the ColorMunki software if running).

Viewing 2 posts - 1 through 2 (of 2 total)

You must be logged in to reply to this topic.