I’d like to run measurement reports of uncalibrated displays (for evaluating out-of-box conditions of these displays for review purposes), but the interface greys out the button unless a calibration / profile is selected. Is there a trick to running a measurement report on displays with no calibration applied?
Thanks in advance,
Allyn
Hi, the next snapshot update will provide that functionality (ETA: in a few days).
I noticed the patch go in very quickly. Thanks for adding this possibility. It will enable the use of DispcalGUI for more display testing purposes!
-
This reply was modified on 2015-10-19 22:07:40 by malventano.
Updated snapshot is now available (3.0.4.5). Via 0install | Standalone
Am I missing something, or has the ability to run measurement reports on uncalibrated displays disappeared again? In 3.1.3.1, the button is greyed out unless a calibration profile is selected. It would be great to have this functionality back for display testing.
Check your settings, you need to select an actual display on the display & instrument tab or enable “use simulation profile as target profile” on the verification tab and set tone curve to “unmodified”.
Thanks, that did the trick.